[esa-t474] Updates
Yury Kolomensky
YGKolomensky at lbl.gov
Wed Nov 7 22:14:27 GMT 2007
Hi Mark and Alex,
I tend to agree with Alex's suggestions. In terms of whether we
understand why 1 and 2 behave so poorly... There is a potential
smoking gun in the plots vs ESA temperature. There is obviously some
correlation, but what could be the cause ? Cables is one obvious
possibility, although the effect seems to be huge for just the cable
explanation. Let's see... Assuming the offset at the time was around
250 um, we are looking at a 5% drift (12/250) per measly 0.3 degree
C. If it's due to an I/Q phase shift, we are talking about 18 degrees
at S-band, or a 5 mm length change. if the cable were a bare copper
waveguide, it would need the length of about 1 km to expand that
much. RF cables usually have better thermal coefficients for both
phase and attenuation, so this is doubtful.
However, there is a caveat. First off, what ESA temperature are you
plotting, Mark ? It should be rack temperature, which is more
representative of the environment. BPM temperatures are presumably
more stable because the BPMs are on a dedicated water chiller. But
even the rack temperature would not be representative of the
temperature *outside* (although it would track it), or of the
temperature of the west wall, along which the cables run. That
presumably drifts a bit more.
Anyway, I think we have the conclusion that over the short distance
scales (triplet) stability is OK, and we are limited by BPMs 1 and 2
to test the long lever arms. What I am wondering though is what you
would get if you just tried to extrapolate from BPMs 9,10 (which are
pretty good) to 3,4,5 ? Or the other way, 3,4,5 (which would give you
a little better handle on angle) to 9 or 10 ? This would have a
disadvantage that any differential drift within a triplet is
amplified, but the advantage is that the environment is the same, so
perhaps we can set a limit on long-term stability of such configuration.
Yury
On Nov 7, 2007, at 9:22 AM, Alexey Lyapin wrote:
>
> Hi Mark et al,
>
> I suggest - for consistency - we show 1 vs 2 (fig23_12) as we do 4
> vs 3,5 and 10 vs 9, 11 to show the intra-station stability. Then,
> we have the linked system stability plots, which don't look as good
> as we hoped. We suspect BPMs 1 and 2, so we show 1 vs 3, 5, 9, 10
> (fig23_10) and 2 vs 3, 5, 9, 10 (fig23_8) and say that both behave
> the same way and the drift is huge, so the screwup is there, but I
> don't think we can really explain it, can we? We could say though
> that we decided to move all the BPM 1,2 electronics to the racks in
> the ESA to avoid additional systematics due to longer cables and
> different environment, to use 14 bit digitisers to improve the
> resolution of 1 and 2, and to build the calibration system and
> monitor drifts online.
>
> Cheers,
> Alex
>
>
> Mark Slater wrote:
>> Dear All,
>> I have tried to address some of the points raised at the meeting
>> on Monday. To that end, I've put some additional analysis points
>> on my elog:
>> http://pcfj.hep.phy.cam.ac.uk:8080/slater/35
>> http://pcfj.hep.phy.cam.ac.uk:8080/slater/36
>> These cover the problems with BPMs 1 and 2 and the position drifts
>> against time. Essentially, it looks like the scale changes in BPM
>> 1 are also present in BPM 2 and these are not based on the cross-
>> talk in this BPM. This therefore explains the relatively poor
>> stability of the linked system. I believe this answers most of the
>> problems that were voiced but do please let me know if this is not
>> the case.
>> I will hopefully add another entry on SVD coefficents, etc. on
>> Friday.
>> Many Thanks for all your patience!
>> Mark
>> P.S. The plots shown in the ELOG also demonstrate some additional
>> runs for the long stability plots.
>> _______________________________________________
>> esa-t474 mailing list
>> esa-t474 at hep.ucl.ac.uk
>> https://mail.hep.ucl.ac.uk/mailman/listinfo/esa-t474
>
>
> _______________________________________________
> esa-t474 mailing list
> esa-t474 at hep.ucl.ac.uk
> https://mail.hep.ucl.ac.uk/mailman/listinfo/esa-t474
More information about the esa-t474
mailing list